[Mristudio-users] ADC-Map Calculation in DTIStudio
Andreas Buchmann
buchmann2 at wisc.edu
Mon Aug 15 18:51:28 EDT 2011
Dear Judy/ Rachael,
ADC is 'apparent diffusion coefficient'. It is defined for the three
coordinate directions, x,y,z (the space in which the subject was in the
scanner).
Note that mean ADC and mean diffusivity is not the same (at least if you
look at the definition). For mean diffusivity, the eigenvalues have to be
estimated from the tensors. In DTI Studio, you can get the 'trace
diffusivity' which equals 3 times the mean diffusivity.
Best
Andreas
_____
From: mristudio-users-bounces at mristudio.org
[mailto:mristudio-users-bounces at mristudio.org] On Behalf Of Rachael Gonzales
Sent: Montag, 15. August 2011 16:32
To: DTI Studio, ROI Editor, DiffeoMap Questions/Support
Subject: [Mristudio-users] ADC-Map Calculation in DTIStudio
Hi DTI experts,
In DTIStudio, under the DTI map tab there is an ADC Map calculation
available. After the calculation, an ADC map - mean image is generated, is
this the apparent or average diffusion coefficient?
Thanks,
Rachael
****************************************************************************
*************************************************
Judy, in DtiStudio, ADC (or also called MD) should be obtained from "Trace".
Trace is defined as Dxx+Dyy+Dzz. If you divide it by 3, it becomes
"Average diffusion constant (ADC)" or "Mean diffusivity (MD)". We like
to use the term "trace" because it has clear mathematical definition
while ADC is also used for "Apparent diffusion constant", which has a
different meaning from "average diffusion constant".
On Mon, Apr 11, 2011 at 11:58 AM, Judy James <jjames2 at umc.edu> wrote:
Is any one in this group aware of how to calculate ADC maps from the
original DWI images in DTI studio - not DTI images, like how to set the
gradient table in such a case?
Any input/help will be highly appreciated.
Thank you,
Judy James "
*********************************************************************
Judy Rose James, Ph.D.
Clinical Physicist, MR Imaging Division
Assistant Professor, Department of Radiology
University of Mississippi Medical Center
Jackson, MS - 39216
Ph: (601) 984-2585; Pager: (601) 929-2442
Email: jjames2 at umc.edu
*********************************************************************
________________________________
From: mristudio-users-bounces at mristudio.org
[mristudio-users-bounces at mristudio.org] On Behalf Of Jacquie Hodge
[Jacquie.Hodge at albertahealthservices.ca]
Sent: Friday, April 08, 2011 11:56 AM
To: mristudio-users at mristudio.org
Subject: [Mristudio-users] Fa values
Good morning everyone,
I am using DTI studio for my project where I am interested in comparing the
left and right CST's in children who have suffereed injury in that area and
after reading many studies on DTI decided to compare average FA values of
the tracts as one of my variables. However, I have come across this questino
from one of my supervisors and cannot really answer it and am hoping for
your help/expertise.
Here is goes:
Is the FA value still a viable measure to look at and use to make a
correlation if you are using FA values to pull the white matter tract? It
seems like if the tract is made from FA values then you should not be able
to use the FA as an outcome variable. This question has come up as I
presented my research idea and I'm not sure how to answer it. Most studies
that use DTI use FA thresholds to pull the tracts and then also the FA
values (eg. mean FA) to make comparisons between the tracts.
I really hope that my question makes sense as I have found it very hard to
word properly.
Thank you very much,
Jacquie
Jacquie Hodge MSc candidate in Neuroscience, BSc
Neuroscience, University of Calgary
Research Assistant, Calgary Pediatric Stroke Program
Alberta Children's Hospital
2888 Shaganappi Trail NW, Calgary, AB, Canada T3B 6A8
Phone: 403-955-7733 Fax: 403-955-2922
jacquie.hodge at albertahealthservices.ca
Program website:
http://www.calgaryhealthregion.ca/programs/stroke/pediatric.htm
________________________________
This message and any attached documents are only for the use of the intended
recipient(s), are confidential and may contain privileged information. Any
unauthorized review, use, retransmission, or other disclosure is strictly
prohibited. If you have received this message in error, please notify the
sender immediately, and then delete the original message. Thank you.
Individuals who have received this information in error or are not
authorized to receive it must promptly return or dispose of the information
and notify the sender. Those individuals are hereby notified that they are
strictly prohibited from reviewing, forwarding, printing, copying,
distributing or using this information in any way.
_______________________________________________
mristudio-users mailing list
mristudio-users at mristudio.org
http://lists.mristudio.org/mailman/listinfo/
Unsubscribe, send a blank email to:
mristudio-users-unsubscribe at mristudio.org
_______________________________________________
mristudio-users mailing list
mristudio-users at mristudio.org
http://lists.mristudio.org/mailman/listinfo/
Unsubscribe, send a blank email to:
mristudio-users-unsubscribe at mristudio.org
On Feb 23, 2011, at 4:35 AM, susumu mori wrote:
Andreas is right.
So called ADC (average diffusion constant, apparent diffusion constant) and
MD (mean diffusivity) are obtained by trace/3 (trace divided by 3), while
the trace is (eigenvalue0+eigenvalue1+eigenvalue2).
also, axial diffusivity is same as eigenvalue0 and radial diffusivity is the
average of eigenvalue1 and 2 [ (eigenvalue1+eigenvalue2)/2].
On Wed, Feb 23, 2011 at 8:31 AM, Andreas Buchmann <buchmann2 at wisc.edu>
wrote:
Dear Yo,
I'm using trace diffusivity, which is strongly related to mean diffusivity
(I think, just multiplied with a constant factor, experts correct me if I'm
wrong).
Andreas
-----Original Message-----
From: mristudio-users-bounces at mristudio.org
[mailto:mristudio-users-bounces at mristudio.org] On Behalf Of
107310 at ib.k.u-tokyo.ac.jp
Sent: Dienstag, 22. Februar 2011 19:36
To: DTI Studio, ROI Editor, DiffeoMap Questions/Support
Subject: Re: [Mristudio-users] radial diffusivity, mean diffusivity
Dear Andreas, Ryuzo and DTI experts
I also try to calculate mean diffusivity.
But in DTI studio, only the radial diffusivity and ADCs can calculate at
the 'Image'-tab with ROI.
Is there any function to measure mean diffusivity?
Please give me any suggestion.
Many thanks.
YO
> Dear Andreas
>
>
>
> Thank you very much for your suggestion. I will try.
>
>
>
> Ryuzo
>
>
>
> From: mristudio-users-bounces at mristudio.org
> [mailto:mristudio-users-bounces at mristudio.org] On Behalf Of Andreas
Buchmann
> Sent: Monday, February 21, 2011 11:03 PM
> To: 'DTI Studio, ROI Editor, DiffeoMap Questions/Support'
> Subject: Re: [Mristudio-users] radial diffusivity, mean diffusivity
>
>
>
> Dear Ryuzo,
>
>
>
> yes, you're on the right track with (1). After the estimation, DTI-
Studio
> appends the images to the list of images you have, and you can see
them
> under the 'Image'-tab and save them with right-click on the image.
>
>
>
> Good luck
>
> Andreas
>
>
>
> _____
>
> From: mristudio-users-bounces at mristudio.org
> [mailto:mristudio-users-bounces at mristudio.org] On Behalf Of hanaie
> Sent: Montag, 21. Februar 2011 07:46
> To: mristudio-users at mristudio.org
> Subject: [Mristudio-users] radial diffusivity, mean diffusivity
>
>
>
> Dear DTI experts
>
>
>
> How can I do to obtain the radial diffusivity and mean diffusivity in
DTI
> studio?
>
>
>
> 1. In the " Dti Map" tab, clicking "Tensor, color Map etc" button,
and
> checking the checkboxes "Radial", "Trace"
>
>
>
> 2. After tensor calculation and fiber tracking, if I click "
statistics"
> button in the "Fiber" tab and
>
> I choose the each image and click "go" button in the another window,
can I
> obtain radial and mean diffusivity?
>
>
>
>
>
> Please give me any suggestion.
>
>
>
> Many thanks.
>
>
>
> Ryuzo
>
>
>
>
> _______________________________________________
> mristudio-users mailing list
> mristudio-users at mristudio.org
> http://lists.mristudio.org/mailman/listinfo/
> Unsubscribe, send a blank email to: mristudio-users-unsubscribe@
mristudio.org <http://mristudio.org/>
>
_______________________________________________
mristudio-users mailing list
mristudio-users at mristudio.org
http://lists.mristudio.org/mailman/listinfo/
Unsubscribe, send a blank email to:
mristudio-users-unsubscribe at mristudio.org
_______________________________________________
mristudio-users mailing list
mristudio-users at mristudio.org
http://lists.mristudio.org/mailman/listinfo/
Unsubscribe, send a blank email to:
mristudio-users-unsubscribe at mristudio.org
_______________________________________________
mristudio-users mailing list
mristudio-users at mristudio.org
http://lists.mristudio.org/mailman/listinfo/
Unsubscribe, send a blank email to:
mristudio-users-unsubscribe at mristudio.org
-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://lists.mristudio.org/pipermail/mristudio-users/attachments/20110815/ecdb453e/attachment-0001.html
More information about the mristudio-users
mailing list