[Mristudio-users] Adjusting Gradients for Oblique Angles

Rajagopalan, Venkateswaran rajagov2 at ccf.org
Fri Dec 10 09:01:54 EST 2010


Dear Jeff Sadino,
 
This is what I did for manual oblique angle correction
 
multiply your gradient matrix (I don't know how many directions you have) if you have 12 directional data then you will have 3*13 (one b=0 values too) with the "Image Orientation patient" matrix 
 
[ 1 0 0 0 .8846 -.4664] rewritten as (these are unit vectors with z direction being 0 0 1 not mentioned in the Image Image orientation patient from Dicom header 
 
[ 1    0     0
  0  0.8846  -0.4664
  0   0      1]
 
3*3
 
Multiply the ablove Image orientation patient with your gradient matrix.
 
Thanks
 
Venkat

________________________________

From: mristudio-users-bounces at mristudio.org on behalf of Jeff Sadino
Sent: Thu 12/9/2010 6:09 PM
To: DTI Studio, ROI Editor, DiffeoMap Questions/Support
Subject: [Mristudio-users] Adjusting Gradients for Oblique Angles


Hello, 

Thank you for the DTI training session in Turtle Bay last week.  It was very helpful for us!

I am wondering how to manually adjust the gradients for an oblique angle.  I am using Siemens 3T Trio Tim VB17.

My default gradient (for one direction) is:
0.861, -0.366, 0.354

After clicking "Rotate gradients if applicable" in DTIStudio 3.0.2 (March2010), the gradient is:
0.861, -0.489, 0.142

I couldn't find exactly which tag in the DICOM header contains the rotation information.  The closest I could find is "ImageOrientationPatient".  The value for this is:
[ 1 0 0 0 .8846 -.4664]

How do I go from the default gradient to the rotated gradient?  Also, can I use this same method on the VB13 and VA25 software?

Thank you very much for your help!
Jeff Sadino

===================================

P Please consider the environment before printing this e-mail

Cleveland Clinic is ranked one of the top hospitals
in America by U.S.News & World Report (2009).  
Visit us online at http://www.clevelandclinic.org for
a complete listing of our services, staff and
locations.


Confidentiality Note:  This message is intended for use
only by the individual or entity to which it is addressed
and may contain information that is privileged,
confidential, and exempt from disclosure under applicable
law.  If the reader of this message is not the intended
recipient or the employee or agent responsible for
delivering the message to the intended recipient, you are
hereby notified that any dissemination, distribution or
copying of this communication is strictly prohibited.  If
you have received this communication in error,  please
contact the sender immediately and destroy the material in
its entirety, whether electronic or hard copy.  Thank you.
-------------- next part --------------
A non-text attachment was scrubbed...
Name: not available
Type: application/ms-tnef
Size: 5561 bytes
Desc: not available
Url : http://lists.mristudio.org/pipermail/mristudio-users/attachments/20101210/d8b90776/attachment-0001.bin 


More information about the mristudio-users mailing list