[Mristudio-users] b-matrix from dicom header and from techs

susumu mori susumu at mri.jhu.edu
Mon Jul 12 19:59:29 EDT 2010


Hi all,

For Siemens users, now a new function to extract a gradient table from
Mosaic file header is available. The extracted gradient table is usually
different from the generic gradient table provided by Siemens from the
following points;

> X/Y/Z gradient signs (+/-) are dynamically changed based on image protocol
(e.g. axial/sagittal/coronal). The extracted tables have correct signs.
> The extracted table is more exact because it includes not only the
diffusion-weighing gradients but also a small contribution of imaging
gradients.

The are some mixed information about whether we need to reorient gradient
tables when oblique imaging is used (whether we need to check "rotate if
applicable" option or not). We are testing it but with the scanner we tested
last time (VB17 operating system), we needed to check to this option.

We will follow up this issue soon but we strongly recommend you to test your
scanner. It can be easily tested by scanning a person twice. Once with a
regular axial and one with a strong oblique (30-45 degree). If you are not
sure about your results, you can send us your results for check.

Susumu

On Mon, Jul 12, 2010 at 9:56 AM, Hangyi Jiang <hjiang at jhmi.edu> wrote:

> hi, Jeff,
>
> if the files are in Mosaic format, DtiDtudio can get the gradients from the
> header.
> below, I attached the email I sent to Venkat for your reference.
>
> regards,
>
> hangyi
>
>
> >>>>>>>>>>>>>>>>>>>>>>>>>>>>
> Hi Venkat,
>
> this is one of the most confusions to deal with DICOM files.
>
> according the DICOM standard and Siemens documentation, the gradients (bvec
> as you mentioned) in DICOM header should be defined in patient space (or
> image matrix space), so the one from dcm2nii should be correct for your
> dataset.
>
> the 1st one (from your MR techs) seems that it is the original definition
> based on scanner's coordinate system. this one should be adjusted
> dynamically during scanning according to patient position and orientation.
>  by the way, we usually named it as gradient table, not b-matrix. b-matrix
> has 6 components for each of DWIs. b-matrix can be calculated from the
> gradient table.
>
> as a matter of fact, Siemens also saves adjusted b-matrix in the DICOM
> header. if you like, you can also use the adjusted b-matrix from the header
> for tensor calculation. in fact, using adjusted b-matrix is recommended by
> Siemens since it also considered other factors (like gradient cross-over
> effects) for image acquisition.
>
> hope it helps.
>
> hangyi
> >>>>>>>>>>>>>>>>>>>>>>>>>>>>
>
>
>
> ________________________________________
> From: mristudio-users-bounces at mristudio.org [
> mristudio-users-bounces at mristudio.org] On Behalf Of Jeff Sadino [
> jsadino.queens at gmail.com]
> Sent: Saturday, July 10, 2010 10:51 PM
> To: DTI Studio, ROI Editor, DiffeoMap Questions/Support; rajagov2 at ccf.org
> Subject: Re: [Mristudio-users] b-matrix from dicom header and from techs
>
> Hi Venkat,
>
> We had the same question on our Siemens VA25A DTIs.  The Siemen's tech
> support recommended a program called MRIConvert (
> http://lcni.uoregon.edu/~jolinda/MRIConvert/<http://lcni.uoregon.edu/%7Ejolinda/MRIConvert/>)
> to extract the gradients.  I don't have our tables in front of me, but the
> .8, 0, and .4 values seem pretty close to what I remember.  Siemens said
> that they suspected the 1, 0, .5 values were in the scanner direction and
> the .8, 0, .4 values were in the patient direction, since I think Siemens
> automatically adjusts for that.  Just be careful when using the MRIConvert
> program to extract gradients, since it flips the y sign for FSL purposes.
>
> Cheers,
> Jeff
>
> On Fri, Jul 9, 2010 at 9:33 AM, Rajagopalan, Venkateswaran <
> rajagov2 at ccf.org<mailto:rajagov2 at ccf.org>> wrote:
>
>
> Dear All,
>
> we collect our data from Siemens scanner. I got b-matrix from our MR techs
> for software version VA25A b matrix is as follows
>
>  0: 0.000, 0.000, 0.000
>  1: 1.000, 0.000, 0.500
>  2: 0.000, 0.500, 1.000
>  3: 0.500, 1.000, 0.000
>  4: 1.000, 0.500, 0.000
>  5: 0.000, 1.000, 0.500
>  6: 0.500, 0.000, 1.000
>  7: 1.000, 0.000, -0.500
>  8: 0.000, -0.500, 1.000
>  9: -0.500, 1.000, 0.000
>  10: 1.000, -0.500, 0.000
>  11: 0.000, 1.000, -0.500
>  12: -0.500, 0.000, 1.000
>
> I used dcm2nii software (to convert hem into nifti file for soem other
> purpose  i also gives me bvec and bval files) which gives me the following
> matrix from the dicom header file
>
> 0       0       0
> 0.89    0       -0.45
> 0       0.45    -0.89
> 0.45    0.89    0
> 0.89    0.45    0
> 0       0.89    -0.45
> 0.45    0       -0.89
> 0.89    0       0.45
> 0       -0.45   -0.89
> -0.45   0.89    0
> 0.89    -0.45   0
> 0       0.89    0.45
> -0.45   0       -0.89
>
> Since not only the values are different but +/- signs are also different in
> DW directions. So I am wondering whether any one could tell me which is the
> correct table to go with.
>
> Thanks
>
> Venkat
>
>
> ===================================
>
> P Please consider the environment before printing this e-mail
>
>
>
>
> Cleveland Clinic is ranked one of the top hospitals
> in America by U.S.News & World Report (2009).
> Visit us online at http://www.clevelandclinic.org<
> http://www.clevelandclinic.org/> for
> a complete listing of our services, staff and
> locations.
>
>
> Confidentiality Note:  This message is intended for use
> only by the individual or entity to which it is addressed
> and may contain information that is privileged,
> confidential, and exempt from disclosure under applicable
> law.  If the reader of this message is not the intended
> recipient or the employee or agent responsible for
> delivering the message to the intended recipient, you are
> hereby notified that any dissemination, distribution or
> copying of this communication is strictly prohibited.  If
> you have received this communication in error,  please
> contact the sender immediately and destroy the material in
> its entirety, whether electronic or hard copy.  Thank you.
>
>
> _______________________________________________
> mristudio-users mailing list
> mristudio-users at mristudio.org<mailto:mristudio-users at mristudio.org>
> http://lists.mristudio.org/mailman/listinfo/
> Unsubscribe, send a blank email to:
> mristudio-users-unsubscribe at mristudio.org<mailto:
> mristudio-users-unsubscribe at mristudio.org>
>
>
>
> _______________________________________________
> mristudio-users mailing list
> mristudio-users at mristudio.org
> http://lists.mristudio.org/mailman/listinfo/
> Unsubscribe, send a blank email to:
> mristudio-users-unsubscribe at mristudio.org
>