Hi Venkat,

We had the same question on our Siemens VA25A DTIs.  The Siemen's tech
support recommended a program called MRIConvert (
http://lcni.uoregon.edu/~jolinda/MRIConvert/) to extract the gradients.  I
don't have our tables in front of me, but the .8, 0, and .4 values seem
pretty close to what I remember.  Siemens said that they suspected the 1, 0,
.5 values were in the scanner direction and the .8, 0, .4 values were in the
patient direction, since I think Siemens automatically adjusts for that.
Just be careful when using the MRIConvert program to extract gradients,
since it flips the y sign for FSL purposes.

Cheers,
Jeff

On Fri, Jul 9, 2010 at 9:33 AM, Rajagopalan, Venkateswaran <rajagov2 at ccf.org
> wrote:

>
> Dear All,
>
> we collect our data from Siemens scanner. I got b-matrix from our MR techs
> for software version VA25A b matrix is as follows
>
>  0: 0.000, 0.000, 0.000
>  1: 1.000, 0.000, 0.500
>  2: 0.000, 0.500, 1.000
>  3: 0.500, 1.000, 0.000
>  4: 1.000, 0.500, 0.000
>  5: 0.000, 1.000, 0.500
>  6: 0.500, 0.000, 1.000
>  7: 1.000, 0.000, -0.500
>  8: 0.000, -0.500, 1.000
>  9: -0.500, 1.000, 0.000
>  10: 1.000, -0.500, 0.000
>  11: 0.000, 1.000, -0.500
>  12: -0.500, 0.000, 1.000
>
> I used dcm2nii software (to convert hem into nifti file for soem other
> purpose  i also gives me bvec and bval files) which gives me the following
> matrix from the dicom header file
>
> 0       0       0
> 0.89    0       -0.45
> 0       0.45    -0.89
> 0.45    0.89    0
> 0.89    0.45    0
> 0       0.89    -0.45
> 0.45    0       -0.89
> 0.89    0       0.45
> 0       -0.45   -0.89
> -0.45   0.89    0
> 0.89    -0.45   0
> 0       0.89    0.45
> -0.45   0       -0.89
>
> Since not only the values are different but +/- signs are also different in
> DW directions. So I am wondering whether any one could tell me which is the
> correct table to go with.
>
> Thanks
>
> Venkat
>
> ===================================
>
> P Please consider the environment before printing this e-mail
>
>
>
> Cleveland Clinic is ranked one of the top hospitals
> in America by U.S.News & World Report (2009).
> Visit us online at http://www.clevelandclinic.org for
> a complete listing of our services, staff and
> locations.
>
>
> Confidentiality Note:  This message is intended for use
> only by the individual or entity to which it is addressed
> and may contain information that is privileged,
> confidential, and exempt from disclosure under applicable
> law.  If the reader of this message is not the intended
> recipient or the employee or agent responsible for
> delivering the message to the intended recipient, you are
> hereby notified that any dissemination, distribution or
> copying of this communication is strictly prohibited.  If
> you have received this communication in error,  please
> contact the sender immediately and destroy the material in
> its entirety, whether electronic or hard copy.  Thank you.
>
>
> _______________________________________________
> mristudio-users mailing list
> mristudio-users at mristudio.org
> http://lists.mristudio.org/mailman/listinfo/
> Unsubscribe, send a blank email to:
> mristudio-users-unsubscribe at mristudio.org
>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://lists.mristudio.org/pipermail/mristudio-users/attachments/20100710/b4d2b306/attachment.html 


More information about the mristudio-users mailing list