[Mristudio-users] Coordinate system for diffusion-gradient directions?
susumu mori
susumu at mri.jhu.edu
Tue Jan 26 14:24:03 EST 2010
Hi Dani,
This issue has been touched upon several times but because it is very
important, I want to re-describe it here.
> All the users who are using the following conditions do not have to worry
anything about gradient coordinate system (magnet frame, patient frame,
imaging frame). Based on the patient position and imaging parameters
(oblique angles), the gradient table is recalculated within the scanner
on-the-fly to assure that gradient table x/y/z correspond to image x/y/z.
When [1, 0, 0] is applied, it is along image X axis.
All GE scanners with all operating systems
Philips scanners with all operating systems except that "Gradient Overplus =
on"
Siemens scanners after VB15 operating system
> The following scanner users have to be worried about gradient coordinates.
When oblique plane is used, the gradient table HAS TO BE recalculated
because the gradient table is in the magnet x/y/z.
Philips scanners when "Gradient Opverplus = on"
Siemens scanners before VB15
If you are using DICOM or MOSAIC, you have to check "Rotate gradients if
applicable" checkbox beneath the gradient table window.
If you are using Philips REC/PAR, please use "DTI_gradient_table _creator"
at http://godzilla.kennedykrieger.org/~jfarrell/software_web.htm written by
Jonathan Farrel.
Susumu
On Mon, Jan 25, 2010 at 11:34 AM, Nanz Daniel Dr. <Daniel.Nanz at usz.ch>wrote:
> Hi
>
> I could not find definitions for the coordinate
> system within which the diffusion directions are to
> be defined for the input into DTI studio?
>
> E.g., on a Philips system, the diffusion direction
> cosines found in the DICOM header (private and public tags)
> are independent of slice orientation as well as of
> patient orientation. Presumably, the direction cosines
> are given in a magnet-fixed coordinate system.
>
> Is this considered and recalculated for tractograpy
> by DTI studio, based on the information in the DICOM
> header (slice orientation in patient coordinate system)?
> Or is the inter-voxel relation of diffusion
> ellipsoid directions neglected altogether by DTI studio?
>
> Or should the directions be re-calculated before
> feeding the data into DTI studio? If so, into what
> coordinate system should the directions be transformed?
>
> With best regards and thanks
> Dani Nanz
>
> _______________________________________________
> Mristudio-users mailing list
> Mristudio-users at mristudio.org
> http://lists.mristudio.org/mailman/listinfo/
> Unsubscribe, send a blank email to:
> Mristudio-users-unsubscribe at mristudio.org
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://lists.mristudio.org/pipermail/mristudio-users/attachments/20100126/be135197/attachment.html
More information about the Mristudio-users
mailing list